Bug: First 1-4 analog samples are 0 and ramping up before data is reliable

Do you have a network connection on your PC?

See the warning:

… looks like you don’t have a calibration file installed for your specific Logic hardware.

You can manually download/install the calibration, following:

If you want a deeper dive into the analog filtering & calibration settings, see this post:

(You can optionally ‘hack’ the calibration file to increase the analog bandwidth / reduce the filtering applied)

However, without any calibration file, you can get strange behavior as described above. In particular, you likely don’t have a mGroupDelay setting that aligns the analog/digital timing for your specific device/filtering settings, nor do you have a mImpulseLength setting that should ‘trim’ the leading analog samples and avoid the behavior you’ve described.